Dose calculation when use range shifter #593
-
I wanna ask if the "matRad_calcParticleDose.m" file has considered that the range shifters change the beam current and energy spectrum before the particle enter the body. |
Beta Was this translation helpful? Give feedback.
Replies: 1 comment 1 reply
-
For the pencil-beam algorithm, not really. The range shifter internally changes the WEPL and adds a correction to the beam broadening before calculating the dose distribution of an individual spot using the tabulated depth dose. Beam current (meaning in [A] or protons/s) is not explicitly modeled, only intensity (i.e. total number of protons per beamlet). Thus, the range & deposited dose will change as it would have through water of equivalent depth (the detailed spectral energy deposition specific to the range shifter material is not modeled). The lateral broadening will assume a more realistic material of the range shifter (I think the default is PMMA). |
Beta Was this translation helpful? Give feedback.
For the pencil-beam algorithm, not really. The range shifter internally changes the WEPL and adds a correction to the beam broadening before calculating the dose distribution of an individual spot using the tabulated depth dose. Beam current (meaning in [A] or protons/s) is not explicitly modeled, only intensity (i.e. total number of protons per beamlet). Thus, the range & deposited dose will change as it would have through water of equivalent depth (the detailed spectral energy deposition specific to the range shifter material is not modeled). The lateral broadening will assume a more realistic material of the range shifter (I think the default is PMMA).